MODEL Aquilion 64
X-RAY TUBE
X-RAY TUBE ANODE
Heat storage, hu (X-ray tube anode) 7500000
Heat dissipation rate, hu/min (X-RAY TUBE) 1,386,000 max
Tube cooling (X-ray tube anode) Oil/air
Tube focal spot, mm (X-ray tube anode) 1.6 x 1.4, 0.9 x 0.8 (IEC standard)
POWER NEEDED 200 VAC, 50/60 Hz, 3-phase
Type (CONFIGURATION) Multislice helical
# of slices (X-ray tube anode) 64
TABLE FEATURES
Range of movement
Vertical, cm (Range of movement) 31-94.4
Longitudinal, cm (Range of movement) 219
Scanning range,cm (Range of movement) 180
Max. patient weight, (precision), kg (Range of movement) 205 (±0.25 mm)
Image storage
Hd capacity, gb (IMAGE STORAGE) 180, 720 raw data, max 3,600 rotations
No. online images (IMAGE STORAGE) 160000
Archive (IMAGE STORAGE) 9.4 GB DVD-RAM
GENERATOR
Output, kw (X-RAY GENERATORS) 60
Kvp range (GENERATOR) 80,100,120, 135
Ma range (IMAGING SYSTEM) 1010-50 in 5 mA steps
DICOM 3.0 COMPATIBLE (POWER NEEDED, VAC) Yes
PERFORMANCE
Minimum interscan time, sec (PERFORMANCE) 0
Dynamic scan pace (PERFORMANCE) 200 scans/100 sec
High-contrast spatial resolution 0% mtf, lp/cm (PERFORMANCE) 18
50% mtf, lp/cm (PERFORMANCE) 8
Low-contrast resolution, mm at % at <=4 rads (PERFORMANCE) 2 at 0.3%
Sound, % at <= 2.5 rads (PERFORMANCE) <0.5
DISPLAY
MONITOR 18" LCD color x 2
Matrixes, pixels (DISPLAY) 1280 x 1024
Range of ct numbers (DISPLAY) -1,536 to +8,191
Image enlarging scale (DISPLAY) Up to 20x
Max # slices displayed simultaneously (DISPLAY) 16
HELICAL SCANNING (DISPLAY) Yes
Max scan time, sec (DISPLAY) 100
Max scan volume, cm (DISPLAY) 175
Spatial resolution, lp/cm (DISPLAY) 18 at 0% MTF
Pitch (DISPLAY) 0.8-96 mm/sec couch-top speed
Reconstruction time per image, sec (DISPLAY) 0.1
FLOOR SPACE NEEDED
Size, m2 (FLOOR SPACE NEEDED) 27 (25 short couch)
SCORING OF CORONARY ARTERY CALCIFICATION (IMAGE STORAGE) Optional
IMAGE PROCESSING (MAXIMUM DISPLAY)
Cpu (NETWORK/DATABASE) 32-bit processor x 2
Scan fovs, cm (IMAGE PROCESSING) 18,24,32,40, 50
Image reconstruction matrixes (IMAGE PROCESSING) 512 x 512
Reconstruction time
Per slice, sec (Reconstruction time) 0.1
For localization scan, sec (Reconstruction time) Real time
GANTRY
Geometry (GANTRY) Rotate-rotate, slip ring, multislice
DETECTOR (SCATTERED LIGHT) Solid-state
Rows (GANTRY) 64
Elements/row (GANTRY) 64 x 896
# detection channels (GANTRY) 64 x 896
Rotation times, sec 360 (GANTRY) 0.5,0.75,1,1.5,2, 3; optional 0.4
Partial (GANTRY) 0.32; optional 0.25
Slice thickness, mm (GANTRY) 0.5 x 64: 1 x 32; 2 x 16; 0.5,1,2,3,4,6, 8 (all x 4)
X-ray fan beam angle, &#194;° (GANTRY) 49.2
Gantry angle deg (GANTRY) ±30
Gantry size, hxwxd,cm (GANTRY) 195 x 233 x 96
Gantry weight, kg (GANTRY) 1750
Gantry opening, cm (GANTRY) 72
Scan localizer (GANTRY) Laser
OTHER ATTRIBUTES (Interference compensation) Cerebral blood-flow analysis syatem; quantitive bone mineral analysis; CT fluoro; full DICOM featue set; pediatric scanning; SureStart contrast tracking; SureScan real-time imaging; volume rendered 3-D; ECG gating; cardiac scoring; cardiac function analys
FDA CLEARANCE (Interference compensation) Yes
CE MARK (MDD) (Interference compensation) Yes
MARKETING REGION (Interference compensation) Worldwide
Our specifications
Год производства: 2006
Serial number: HCA06Z2451
Tube model: MSC7078DREV2
Serial number: 92421-R4 (replaced 25/08/2014)
Tube scan seconds: 362.446
Gantry count: 2.129.181
Software version: V5.10*R007
Including: Mallinckrodt Optivantage DH contrast injector